[2026 update] Charge states for a 4 kDa peptide, worked through posts 31–60

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

AK
a.kravchenkoTL2 Moderator25 Mar 2026#31
g.valckenaere, post #8: Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Coming back to post #29, because the follow-up matters more than the original answer.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

8 likes in reply to #8 4mo
B
BirkelandTL3Regular26 Mar 2026 · edited#32
d.bramley, post #1: On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a monoisotopic mass close to 4113.6 Da. On an electrospray instrument I would expect to see the multiply charged series rather than… Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

2 likes in reply to #1 4mo
VR
v.rautioTL2 Moderator26 Mar 2026#33

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 4mo
CT
cannula_traceTL3Regular27 Mar 2026#34

post #33 answers the question as asked. The question underneath it is different.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

27 likes 4mo
GA
g.amankwahTL2 Moderator27 Mar 2026#35
a.pereira, post #12: post #11 answers the question as asked. The question underneath it is different. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of… Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

5 likes in reply to #12 4mo
OF
outline_firstTL3Wiki editor28 Mar 2026#36

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes 4mo
CC
c.castellanosTL2 Moderator28 Mar 2026#37

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 4mo
NE
n.ekstromTL2Regular29 Mar 2026#38

post #37 is right about the mechanism and I think understates the practical bit.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

20 likes 4mo
SV
s.vukovicTL2 Moderator29 Mar 2026 · edited#39
ppm_error, post #17: I read post #15 twice before replying, because I had assumed the opposite. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

19 likes in reply to #17 4mo
SS
steady_stateTL3Regular29 Mar 2026#40

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

8 likes 4mo
GV
g.valckenaereTL3Regular30 Mar 2026#41

This follows post #38 rather than contradicting it.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

4 likes 4mo
BC
b.correiaTL2 Moderator30 Mar 2026#42
n.villalobos, post #23: This follows post #20 rather than contradicting it. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

13 likes in reply to #23 4mo
H
HHidalgoTL2Member31 Mar 2026#43
c.castellanos, post #37: Resolution: "high resolution" commonly means Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes in reply to #37 4mo
SD
st.dialloTL2 Moderator31 Mar 2026#44

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 4mo
K
KAnderssonTL3Regular31 Mar 2026#45

Picking up post #42: that is the part I would want checked first.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

7 likes 4mo
ST
s.teixeiraTL2 Moderator1 Apr 2026#46
h.iyer, post #21: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Coming back to post #44, because the follow-up matters more than the original answer.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

18 likes in reply to #21 4mo
DS
d.szymanskiTL3Wiki editor1 Apr 2026#47

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 4mo
EN
e.ndiayeTL2 Moderator2 Apr 2026 · edited#48

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

1 like 4mo
VD
vial_deskTL3Regular2 Apr 2026#49

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

12 likes 4mo
EM
e.mensaTL2 Moderator2 Apr 2026#50

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

25 likes 4mo
BP
baseline_peakTL2Member3 Apr 2026 · edited#51

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 4mo
BJ
b.jansenTL2 Moderator3 Apr 2026#52

post #51 is right about the mechanism and I think understates the practical bit.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

29 likes 4mo
MS
m.stephanopoulosTL3Regular4 Apr 2026#53
cannula_trace, post #34: post #33 answers the question as asked. The question underneath it is different. Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

I read post #51 twice before replying, because I had assumed the opposite.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

10 likes in reply to #34 4mo
JS
j.silvaTL2 Moderator4 Apr 2026#54
n.villalobos, post #23: This follows post #20 rather than contradicting it. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

3 likes in reply to #23 4mo
RG
r.girardTL24 Apr 2026#55
ZI
z.iyerTL2 Moderator5 Apr 2026#56

post #55 answers the question as asked. The question underneath it is different.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

22 likes 4mo
DB
d.bakkerTL2 Moderator5 Apr 2026#57

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

6 likes 4mo
CS
c.silvaTL2 Moderator6 Apr 2026#58
a.kravchenko, post #31: Coming back to post #29, because the follow-up matters more than the original answer. Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

1 like in reply to #31 4mo
HN
h.nicolaidesTL3Regular6 Apr 2026#59

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

31 likes 4mo
KK
k.karlsenTL2 Moderator6 Apr 2026 · edited#60
forest_plot, post #11: On post #7 — agreed on the reasoning, with one qualification. Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

15 likes in reply to #11 4mo