sourced_claimsTL3Regular8 Feb 2026#31

Coming back to post #29, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

18 likes 6mo
SG
s.grimaldiTL2 Moderator9 Feb 2026#32
a.kwiatkowski, post #8: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Picking up post #29: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

7 likes in reply to #8 6mo
DV
dr.villanuevaTL3Physician10 Feb 2026 · edited#33

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

1 like 6mo
EI
e.iyerTL2 Moderator11 Feb 2026#34

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 5mo
MH
ms_hollowayTL4Mass spectrometrist12 Feb 2026#35

I read post #33 twice before replying, because I had assumed the opposite.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

12 likes 5mo
RS
r.serranoTL2 Moderator13 Feb 2026#36
k.haddad, post #12: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

4 likes in reply to #12 5mo
OB
owen.bradyTL4 Moderator14 Feb 2026#37
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes 5mo
KD
k.dahlbergTL2 Moderator15 Feb 2026#38

post #37 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 5mo
WT
week_threeTL1Member16 Feb 2026#39

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

8 likes 5mo
JI
j.ivaturiTL2 Moderator17 Feb 2026#40

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

2 likes 5mo
YA
y.adebayoTL2 Moderator18 Feb 2026#41

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

32 likes 5mo
DV
dr.villanuevaTL3Physician19 Feb 2026#42

I read post #40 twice before replying, because I had assumed the opposite.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 5mo
NS
n.silvaTL2 Moderator20 Feb 2026#43

post #42 is right about the mechanism and I think understates the practical bit.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

6 likes 5mo
MH
ms_hollowayTL4Mass spectrometrist21 Feb 2026 · edited#44
Knowlton, post #28: Coming back to post #26, because the follow-up matters more than the original answer. Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

16 likes in reply to #28 5mo
LS
l.salinasTL2 Moderator22 Feb 2026#45

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 5mo
SL
s.leclercTL4 Moderator23 Feb 2026#46

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

1 like 5mo
JS
j.steinerTL2 Moderator23 Feb 2026#47

post #46 answers the question as asked. The question underneath it is different.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

10 likes 5mo
AR
a.reyesTL4 Admin24 Feb 2026#48
IRenaudin, post #10: Coming back to post #8, because the follow-up matters more than the original answer. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

On post #44 — agreed on the reasoning, with one qualification.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

23 likes in reply to #10 5mo
HM
h.mbekiTL225 Feb 2026#49
KR
k.radichTL2 Moderator26 Feb 2026#50

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes 5mo
NT
n.torrenceTL3Regular27 Feb 2026#51

Worth separating two things that post #47 runs together.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

22 likes 5mo
MA
m.agyemanTL2 Moderator28 Feb 2026#52

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

10 likes 5mo
BM
buffer_marginTL3Regular1 Mar 2026#53

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

1 like 5mo
KA
k.agyemanTL2 Moderator2 Mar 2026#54
s.vanhecke, post #16: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes in reply to #16 5mo
P
PSkarbekTL3Regular3 Mar 2026#55
s.chowdhury, post #9: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

16 likes in reply to #9 5mo
BR
b.restrepoTL2 Moderator3 Mar 2026 · edited#56

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

6 likes 5mo
DP
d.petrescuTL24 Mar 2026#57
TA
t.abubakarTL2 Moderator5 Mar 2026#58

Picking up post #55: that is the part I would want checked first.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

31 likes 5mo
CD
c.dahlbergTL2 Moderator6 Mar 2026#59
k.radich, post #50: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes in reply to #50 5mo
IC
i.coelhoTL2 Moderator7 Mar 2026#60

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

21 likes 5mo