Isotope patterns and why the monoisotopic peak is not the tallest

EK
ew.kuuselaTL2 Moderator25 Oct 2024#1

Isotope patterns and why the monoisotopic peak is not the tallest — setting out what I have, and where I think it stops being reliable.

Working through the identity arithmetic and I would like it checked.

retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4731.3 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

20 likes 21mo
FH
f.haddadTL2 Moderator8 Nov 2024#2

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

3 likes 21mo
PN
p.novotnyTL2Regular18 Nov 2024#3
f.haddad, post #2: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Worth separating two things that the opening post runs together.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes in reply to #2 20mo
MB
m.balogunTL2 Moderator27 Nov 2024#4

post #3 is right about the mechanism and I think understates the practical bit.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

32 likes 20mo
UC
unit_conversionTL3Regular5 Dec 2024#5

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

16 likes 20mo
IL
i.lehtinenTL2 Moderator13 Dec 2024#6
unit_conversion, post #5: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Picking up post #3: that is the part I would want checked first.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

6 likes in reply to #5 19mo
QZ
q.zhao_qaTL3Quality assurance21 Dec 2024 · edited#7
m.balogun, post #4: post #3 is right about the mechanism and I think understates the practical bit. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

On post #3 — agreed on the reasoning, with one qualification.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

1 like in reply to #4 19mo
ES
e.steinerTL2 Moderator28 Dec 2024#8

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 19mo
VS
vial_slopeTL3Regular4 Jan 2025#9

I read post #7 twice before replying, because I had assumed the opposite.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

23 likes 19mo
SD
s.demirTL2 Moderator11 Jan 2025#10
q.zhao_qa, post #7: On post #3 — agreed on the reasoning, with one qualification. Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

10 likes in reply to #7 19mo
PW
PharmNotes_WhitfieldTL4Pharmacist18 Jan 2025#11

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

16 likes 18mo
NK
n.kuuselaTL224 Jan 2025#12
DO
dr_okonkwoTL4 Moderator31 Jan 2025#13
f.haddad, post #2: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

post #12 is right about the mechanism and I think understates the practical bit.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #2 18mo
JF
j.fonsecaTL2 Moderator6 Feb 2025#14
unit_conversion, post #5: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Worth separating two things that post #10 runs together.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

3 likes in reply to #5 18mo
DS
dr_seongTL3Physician12 Feb 2025#15

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

22 likes 17mo
RF
ro.friskTL2 Moderator18 Feb 2025#16

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes 17mo
OO
orbitrap_olaTL3Mass spectrometrist25 Feb 2025#17
ew.kuusela, post #1: Isotope patterns and why the monoisotopic peak is not the tallest — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked. retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged… Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

1 like in reply to #1 17mo

Suggested topics

TopicParticipantsRepliesViewsActivity
[2026 update] Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a…
DBBIATNAL+116 135 31k 3mo
MALDI versus ESI for a quick identity check
On the subject in the title: MALDI versus ESI for a quick identity check Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
RMTVO 2 2k 11mo
Reading an MS report that only gives you a single number — the long version
Reading an MS report that only gives you a single number — the long version — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I…
TTRAK 2 47k 18mo
Coming back to: Deconvolution of a multiply charged envelope
On the subject in the title: Deconvolution of a multiply charged envelope Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
VRCVEL 2 12k 22mo
Oxidation and deamidation mass shifts, tabulated
Posting this under the heading it deserves: Oxidation and deamidation mass shifts, tabulated Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
AFJVIATNCS+67 76 21k 16mo

Related topics — sharing the tags LC-MS, MALDI-TOF, mass error (ppm)

TopicParticipantsRepliesViewsActivity
[2026 update] Buffers, pH and why a formulation is a formulation
Posting this under the heading it deserves: Buffers, pH and why a formulation is a formulation Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
PDRRS 2 773 3mo
Endotoxin and why it is a separate question from purity
Endotoxin and why it is a separate question from purity Writing it up because I had to work it out twice and would rather nobody else did. I would like to understand what this number means before I repeat it…
DNSGMB 2 22k 4mo
Distinguishing a deletion sequence by mass alone — the long version
On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion. I would like to understand what this number means before I repeat it…
TFVKPFIDLF+38 44 60k 11mo
Second pass at: Isotonicity and what happens when you get it wrong
Second pass at: Isotonicity and what happens when you get it wrong — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked.…
NSFNCATHRE+1 5 2.6k 11mo
High-resolution MS on aggregates and what it can see — one year on
High-resolution MS on aggregates and what it can see — one year on Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
NGNDBEIWCO+54 58 1.2k 23mo