What a 2 ppm match establishes, and what it does not posts 31–60

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

RH
revision_historyTL3Wiki editor6 Jul 2025#31

Worth separating two things that post #27 runs together.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

1 like 13mo
AA
a.adeyemiTL2 Moderator10 Jul 2025#32

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 13mo
WT
week_threeTL1Member14 Jul 2025#33
s.vogel, post #7: Picking up post #4: that is the part I would want checked first. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

17 likes in reply to #7 12mo
JI
j.ivaturiTL2 Moderator17 Jul 2025 · edited#34

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

7 likes 12mo
HO
h.oyelowoTL2Regular21 Jul 2025#35

On post #31 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 12mo
RZ
ro.zielinskiTL2 Moderator25 Jul 2025#36
BDraganov, post #16: Picking up post #13: that is the part I would want checked first. Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

post #35 answers the question as asked. The question underneath it is different.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes in reply to #16 12mo
M
microgramsTL228 Jul 2025#37
MR
m.radichTL2 Moderator1 Aug 2025#38

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

4 likes 12mo
DV
dr.villanuevaTL3Physician5 Aug 2025#39

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

6 likes 12mo
EI
e.iyerTL2 Moderator8 Aug 2025#40

post #39 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

1 like 12mo
FF
f.fenwickTL3Regular12 Aug 2025#41

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

2 likes 12mo
LA
l.aguirreTL2 Moderator15 Aug 2025#42
micrograms, post #37: Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

9 likes in reply to #37 11mo
RF
r.friskTL219 Aug 2025#43
HF
h.ferrariTL2 Moderator22 Aug 2025#44

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 11mo
EL
e.lehtinenTL2 Moderator26 Aug 2025 · edited#45

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

1 like 11mo
BA
b.aaltoTL2 Moderator29 Aug 2025#46

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

5 likes 11mo
KF
k.fonsecaTL2 Moderator2 Sep 2025#47
j.vandermolen, post #1: What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it checked. semaglutide has a monoisotopic mass close to 4113.6 Da. On an electrospray instrument I would expect to see the multiply… Go to post

Picking up post #44: that is the part I would want checked first.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

15 likes in reply to #1 11mo
KV
k.vanheckeTL2 Moderator5 Sep 2025#48

Coming back to post #46, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

30 likes 11mo
W
WoodhouseTL2Member9 Sep 2025#49

post #48 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 11mo
SR
s.radichTL2 Moderator12 Sep 2025#50
ar.kravchenko, post #5: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Worth separating two things that post #46 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

3 likes in reply to #5 10mo
RP
r.petrovTL2 Moderator16 Sep 2025#51

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

8 likes 10mo
PE
ppm_errorTL3Analytical chemist19 Sep 2025#52

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

2 likes 10mo
BV
b.vanheckeTL2 Moderator22 Sep 2025#53
ar.kravchenko, post #5: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

On post #49 — agreed on the reasoning, with one qualification.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #5 10mo
C
chromatogramTL4Analytical chemist26 Sep 2025#54

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

26 likes 10mo
NV
n.vukovicTL2 Moderator29 Sep 2025#55

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

5 likes 10mo
AD
appeals_deskTL3Regular2 Oct 2025#56

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 10mo
YR
y.rahimiTL2 Moderator6 Oct 2025#57
appeals_desk, post #56: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Worth separating two things that post #53 runs together.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes in reply to #56 10mo
P
preregisteredTL3Research methods9 Oct 2025#58

post #57 is right about the mechanism and I think understates the practical bit.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

19 likes 10mo
NS
n.stanescuTL2 Moderator12 Oct 2025#59

Coming back to post #57, because the follow-up matters more than the original answer.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

2 likes 10mo
LC
lu.cabreraTL2 Moderator16 Oct 2025#60

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes 9mo