MR
m.ramosTL2 Moderator27 Jul 2025#31
l.wikstrom, post #27: Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

Worth separating two things that post #27 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

21 likes in reply to #27 12mo
CB
careful_beginnerTL1Member29 Jul 2025#32

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

9 likes 12mo
SR
s.roosTL2 Moderator30 Jul 2025 · edited#33

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

1 like 12mo
DN
desiccant_notesTL2Member1 Aug 2025#34

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 12mo
EM
e.mbekiTL2 Moderator2 Aug 2025#35

On post #31 — agreed on the reasoning, with one qualification.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

15 likes 12mo
GT
g.tanakaTL3Regular3 Aug 2025#36

post #35 answers the question as asked. The question underneath it is different.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

6 likes 12mo
MI
m.ilungaTL2 Moderator5 Aug 2025#37

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes 12mo
RH
revision_historyTL3Wiki editor6 Aug 2025#38
m.ilunga, post #37: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

30 likes in reply to #37 12mo
GE
g.ekstromTL2 Moderator8 Aug 2025#39

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

10 likes 12mo
L
LeitermanTL3Regular9 Aug 2025#40

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

3 likes 12mo
HE
h.eriksenTL2 Moderator10 Aug 2025#41
revision_history, post #38: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

15 likes in reply to #38 12mo
CL
c.lundgrenTL2 Moderator12 Aug 2025#42

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

30 likes 12mo
NN
n.nybergTL2 Moderator13 Aug 2025#43

This follows post #40 rather than contradicting it.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 11mo
TP
t.pereiraTL2 Moderator15 Aug 2025#44
l.wikstrom, post #27: Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

3 likes in reply to #27 11mo
FV
f.villalobosTL2 Moderator16 Aug 2025#45
m.ilunga, post #37: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

10 likes in reply to #37 11mo
CG
c.grimaldiTL2 Moderator17 Aug 2025#46

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

23 likes 11mo
DO
dr_okonkwoTL419 Aug 2025#47
JF
j.fonsecaTL2 Moderator20 Aug 2025#48

Coming back to post #46, because the follow-up matters more than the original answer.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

1 like 11mo
PW
PharmNotes_WhitfieldTL4Pharmacist21 Aug 2025#49
e.mbeki, post #35: On post #31 — agreed on the reasoning, with one qualification. Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples. Go to post

post #48 is right about the mechanism and I think understates the practical bit.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

29 likes in reply to #35 11mo
NK
n.kuuselaTL2 Moderator23 Aug 2025#50
dr_okonkwo, post #47: Picking up post #44: that is the part I would want checked first. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes in reply to #47 11mo
AD
appeals_deskTL3Regular24 Aug 2025#51
h.eriksen, post #41: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

9 likes in reply to #41 11mo
HL
h.lindqvistTL2 Moderator25 Aug 2025#52

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

2 likes 11mo
TH
TL4_HalvorsenTL4Leader · Journal club27 Aug 2025#53

On post #49 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 11mo
JV
j.vogelTL2 Moderator28 Aug 2025#54
GDashwood, post #8: Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

29 likes in reply to #8 11mo
MH
m.haddadTL2Regular29 Aug 2025#55
revision_history, post #38: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

5 likes in reply to #38 11mo
AD
a.delgadoTL2 Moderator31 Aug 2025#56

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

1 like 11mo
LI
l.ibarraTL2Regular1 Sep 2025 · edited#57

Worth separating two things that post #53 runs together.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 11mo
AK
a.kirchnerTL2 Moderator2 Sep 2025#58

post #57 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

21 likes 11mo
AL
aliquot_lineTL3Regular3 Sep 2025#59

Coming back to post #57, because the follow-up matters more than the original answer.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

20 likes 11mo
EN
e.nilsenTL2 Moderator5 Sep 2025#60

Picking up post #57: that is the part I would want checked first.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

9 likes 11mo