Second pass at: What a 2 ppm match establishes, and what it does not

BS
buffer_sheetTL3Regular4 Apr 2026#1

Second pass at: What a 2 ppm match establishes, and what it does not — setting out what I have, and where I think it stops being reliable.

Working through the identity arithmetic and I would like it checked.

semaglutide has a monoisotopic mass close to 4113.6 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4113.6 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

0 likes 4mo
NK
n.kaufmannTL2 Moderator10 Apr 2026#2

This follows the opening post rather than contradicting it.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

31 likes 4mo
ST
stopper_traceTL2Member14 Apr 2026#3

Worth separating two things that the opening post runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

16 likes 3mo
MG
m.guerreroTL2 Moderator17 Apr 2026#4

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

6 likes 3mo
MM
methods_marginTL3Regular20 Apr 2026#5
n.kaufmann, post #2: This follows the opening post rather than contradicting it. For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

Coming back to post #3, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

1 like in reply to #2 3mo
KB
k.batistaTL2 Moderator23 Apr 2026 · edited#6

Picking up post #3: that is the part I would want checked first.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes 3mo
CR
crossover_reviewTL3Regular26 Apr 2026#7

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

22 likes 3mo
NB
n.boatengTL2 Moderator29 Apr 2026#8

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

10 likes 3mo
F
FFaulknerTL3Regular2 May 2026#9

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

3 likes 3mo
VO
v.okonkwoTL2 Moderator4 May 2026 · edited#10

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 3mo
BE
bench_entryTL3Regular7 May 2026#11

This follows post #8 rather than contradicting it.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes 3mo
BW
b.wikstromTL2 Moderator10 May 2026#12
FFaulkner, post #9: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

I read post #10 twice before replying, because I had assumed the opposite.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

2 likes in reply to #9 3mo
BJ
b.jankowiakTL3Regular12 May 2026#13

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

8 likes 3mo
IW
i.wojcikTL214 May 2026#14
IL
integrator_logTL3Regular17 May 2026#15
k.batista, post #6: Picking up post #3: that is the part I would want checked first. I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the… Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes in reply to #6 2mo
SS
s.salgadoTL2 Moderator19 May 2026#16

Coming back to post #14, because the follow-up matters more than the original answer.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

4 likes 2mo
BS
buffer_sheetTL3Regular22 May 2026#17

post #16 answers the question as asked. The question underneath it is different.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

13 likes 2mo
FL
f.lindholmTL2 Moderator24 May 2026 · edited#18
m.guerrero, post #4: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

27 likes in reply to #4 2mo
D
DOdendaalTL326 May 2026#19
CM
c.marchettiTL2 Moderator28 May 2026#20

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

8 likes 2mo
ZV
z.vogelTL2 Moderator31 May 2026#21

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

10 likes 2mo
W
WoodhouseTL2Member2 Jun 2026#22
methods_margin, post #5: Coming back to post #3, because the follow-up matters more than the original answer. Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

3 likes in reply to #5 2mo
CV
ca.vermeulenTL2 Moderator4 Jun 2026#23

Worth separating two things that post #19 runs together.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes 2mo
GF
gradient_fileTL2Member6 Jun 2026#24

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

22 likes 2mo
SK
s.kravchenkoTL2 Moderator8 Jun 2026 · edited#25

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

6 likes 2mo
EL
endpoint_lineTL3Regular10 Jun 2026#26
m.guerrero, post #4: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

1 like in reply to #4 2mo
ZA
z.adeyemiTL2 Moderator12 Jun 2026#27

On post #23 — agreed on the reasoning, with one qualification.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

31 likes 2mo
R
RidgewayTL3Regular14 Jun 2026#28

post #27 answers the question as asked. The question underneath it is different.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

15 likes 1mo
IG
i.grimaldiTL2 Moderator16 Jun 2026#29

I read post #27 twice before replying, because I had assumed the opposite.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

21 likes 1mo
EF
erratum_fileTL3Regular18 Jun 2026#30

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

9 likes 1mo