Charge states for a 4 kDa peptide, worked through posts 91–120

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

BO
b.oylerTL1Member13 Mar 2025#91

On post #87 — agreed on the reasoning, with one qualification.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

7 likes 17mo
JI
j.ivaturiTL2 Moderator13 Mar 2025#92
a.aguirre, post #52: Resolution: "high resolution" commonly means Go to post

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

1 like in reply to #52 17mo
DN
desiccant_notesTL2Member13 Mar 2025#93

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes 17mo
MR
m.ramosTL2 Moderator13 Mar 2025#94

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

18 likes 17mo
M
microgramsTL2Regular13 Mar 2025#95
KAndersson, post #65: post #64 is right about the mechanism and I think understates the practical bit. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Worth separating two things that post #91 runs together.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

12 likes in reply to #65 17mo
MK
m.kjaerTL213 Mar 2025#96
RH
revision_historyTL3Wiki editor13 Mar 2025#97

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 17mo
EM
e.mbekiTL2 Moderator13 Mar 2025#98

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

25 likes 16mo
SC
sourced_claimsTL3Regular14 Mar 2025#99
n.ibarra, post #58: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

17 likes in reply to #58 16mo
MV
m.vukovicTL2 Moderator14 Mar 2025#100
Makinen, post #36: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

post #99 answers the question as asked. The question underneath it is different.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

7 likes in reply to #36 16mo
AD
ambient_draftTL3Regular14 Mar 2025#101

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

1 like 16mo
AK
ak.kravchenkoTL2 Moderator14 Mar 2025#102
n.ibarra, post #58: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes in reply to #58 16mo
TS
t.steenkampTL2Member14 Mar 2025#103

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

17 likes 16mo
AW
ai.wikstromTL2 Moderator14 Mar 2025#104

post #103 answers the question as asked. The question underneath it is different.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

6 likes 16mo
P
PWendelboeTL1Member14 Mar 2025#105

I read post #103 twice before replying, because I had assumed the opposite.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 16mo
CN
c.nybergTL2 Moderator14 Mar 2025#106
careful_beginner, post #72: This follows post #69 rather than contradicting it. Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

32 likes in reply to #72 16mo
LS
l.sarkissianTL2Member14 Mar 2025#107

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

11 likes 16mo
TM
t.marchettiTL2 Moderator14 Mar 2025#108

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

3 likes 16mo
I
IHollingworthTL2Member15 Mar 2025#109
j.petrov, post #1: On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than… Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes in reply to #1 16mo
NR
n.ramosTL2 Moderator15 Mar 2025#110
KAndersson, post #86: Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

Picking up post #107: that is the part I would want checked first.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

24 likes in reply to #86 16mo
IA
i.aranda_esTL2Translator · ES15 Mar 2025#111

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

9 likes 16mo
RW
r.weissTL2 Moderator15 Mar 2025#112

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

21 likes 16mo
SG
s.grigorescuTL215 Mar 2025#113
WV
w.verhoevenTL2 Moderator15 Mar 2025#114
Leiterman, post #80: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

I read post #112 twice before replying, because I had assumed the opposite.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

2 likes in reply to #80 16mo
WP
weekly_pinTL2Regular15 Mar 2025#115

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

14 likes 16mo
LV
l.vukovicTL2 Moderator15 Mar 2025#116

On post #112 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

28 likes 16mo
SL
sleep_logTL2Regular15 Mar 2025#117
FFaulkner, post #44: Worth separating two things that post #40 runs together. Resolution: "high resolution" commonly means Go to post

Picking up post #114: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes in reply to #44 16mo
II
i.ilungaTL2 Moderator15 Mar 2025#118

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

5 likes 16mo
TH
TL4_HalvorsenTL4Leader · Journal club16 Mar 2025#119

post #118 is right about the mechanism and I think understates the practical bit.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

20 likes 16mo
CA
c.amankwahTL2 Moderator16 Mar 2025#120

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.