Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable.

Working through the identity arithmetic and I would like it checked.

retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4731.3 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

36 likes 16mo
SL
s.lundgrenTL2 Moderator8 Apr 2025#2

Picking up the opening post: that is the part I would want checked first.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

8 likes 16mo
AD
ambient_draftTL3Regular16 Apr 2025#3

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

2 likes 15mo
AK
ak.kravchenkoTL2 Moderator22 Apr 2025#4

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes 15mo
TI
trough_indexTL3Regular29 Apr 2025#5

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

13 likes 15mo
AN
a.norgaardTL2 Moderator5 May 2025#6
ambient_draft, post #3: Resolution: "high resolution" commonly means Go to post

This follows post #3 rather than contradicting it.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

5 likes in reply to #3 15mo
NB
n.bridgewaterTL2Member10 May 2025 · edited#7
h.ferrari, post #1: Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked. retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than… Go to post

Worth separating two things that post #3 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes in reply to #1 15mo
NL
ne.laurentTL2 Moderator16 May 2025#8

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 14mo
B
BDraganovTL2Member21 May 2025#9
ambient_draft, post #3: Resolution: "high resolution" commonly means Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes in reply to #3 14mo
TM
t.marchettiTL2 Moderator26 May 2025#10
s.lundgren, post #2: Picking up the opening post: that is the part I would want checked first. I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because… Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

18 likes in reply to #2 14mo
YA
y.adeyemiTL2 Moderator31 May 2025#11
h.ferrari, post #1: Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked. retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than… Go to post

Picking up post #8: that is the part I would want checked first.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

24 likes in reply to #1 14mo
AT
a.thorneTL2Wiki editor5 Jun 2025#12

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 14mo
GR
g.radichTL210 Jun 2025#13
JR
j.rasmussenTL2Regular15 Jun 2025 · edited#14

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

7 likes 13mo
CC
c.chowdhuryTL2 Moderator19 Jun 2025#15
ambient_draft, post #3: Resolution: "high resolution" commonly means Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

17 likes in reply to #3 13mo
TY
two_year_lineTL3Regular24 Jun 2025#16
a.norgaard, post #6: This follows post #3 rather than contradicting it. Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

33 likes in reply to #6 13mo
DF
d.ferreiraTL2 Moderator29 Jun 2025#17

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 13mo
B
batchlogTL33 Jul 2025#18
IA
id.almeidaTL2 Moderator7 Jul 2025#19

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

11 likes 13mo
BV
bias_varianceTL4Biostatistician12 Jul 2025#20
t.marchetti, post #10: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Coming back to post #18, because the follow-up matters more than the original answer.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

25 likes in reply to #10 13mo
JS
j.steinerTL2 Moderator16 Jul 2025#21

Coming back to post #19, because the follow-up matters more than the original answer.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

22 likes 12mo
KR
k.radichTL2 Moderator20 Jul 2025#22

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

9 likes 12mo
HM
h.mbekiTL2 Moderator25 Jul 2025#23

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

1 like 12mo
BS
b.solbergTL2 Moderator29 Jul 2025#24
ak.kravchenko, post #4: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

post #23 answers the question as asked. The question underneath it is different.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes in reply to #4 12mo
NS
n.silvaTL2 Moderator2 Aug 2025#25

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

29 likes 12mo
SL
s.leclercTL4 Moderator6 Aug 2025#26
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

14 likes 12mo
LS
l.salinasTL2 Moderator10 Aug 2025#27

Worth separating two things that post #23 runs together.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

2 likes 12mo
AR
a.reyesTL4 Admin14 Aug 2025#28
ne.laurent, post #8: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

post #27 is right about the mechanism and I think understates the practical bit.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes in reply to #8 11mo
CD
cannula_driftTL318 Aug 2025#29
SV
sa.vogelTL2 Moderator22 Aug 2025 · edited#30

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

21 likes 11mo