Relative retention time as a better inter-laboratory comparator — does this still hold?

Solved Wiki
Solved by n.broberg in post #6
Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

Jump to the accepted answer →

EA
e.almeidaTL2Member14 Jun 2024#1

Relative retention time as a better inter-laboratory comparator — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below.

A documentation question rather than an analytical one.

I have a certificate in front of me that reports a purity figure, names a technique, gives a wavelength, and stops. No gradient, no column, no injection volume, no chromatogram.

What can I legitimately conclude from that document? My instinct is "almost nothing, but not literally nothing", and I would like to know where the people who read these professionally draw the line.

0 likes 2.1y
RF
ro.friskTL2 Moderator1 Jul 2024#2

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

26 likes 2.1y
WN
w.novakTL3Regular13 Jul 2024#3

Coming back to the opening post, because the follow-up matters more than the original answer.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

8 likes 2y
FW
f.weissTL2 Moderator23 Jul 2024#4
Community wiki post. Any member at trust level 3 or above can edit this post; every edit is recorded. Last edited by r.aldana_pharmd on 27 Oct 2024.
  • 28 Sep 2024 — TL4_Halvorsen: Replaced an unsourced figure with the published one and cited it.
  • 27 Oct 2024 — r.aldana_pharmd: Removed a claim that the cited source did not support.
Editors: TL4_Halvorsen, r.aldana_pharmd

Picking up post #2: that is the part I would want checked first.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

2 likes 2y
OO
orbitrap_olaTL3Mass spectrometrist2 Aug 2024#5

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 2y
NB
n.brobergTL2 Moderator Solution11 Aug 2024#6
orbitrap_ola, post #5: Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

7 likes in reply to #5 2y
DS
dr_seongTL3Physician20 Aug 2024#7

I read post #5 twice before replying, because I had assumed the opposite.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

12 likes 23mo
IA
i.almeidaTL2 Moderator29 Aug 2024#8

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

4 likes 23mo
BJ
b.jankowiakTL3Regular6 Sep 2024#9

On post #5 — agreed on the reasoning, with one qualification.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

27 likes 23mo
JF
j.falkTL2 Moderator14 Sep 2024#10
i.almeida, post #8: Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount. Go to post

post #9 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

13 likes in reply to #8 22mo
EO
e.okaforTL2 Moderator22 Sep 2024#11

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

6 likes 22mo
LT
l.trevinoTL2 Moderator30 Sep 2024#12
e.almeida, post #1: Relative retention time as a better inter-laboratory comparator — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below. A documentation question rather than an analytical one. I have a certificate in front of me that reports a purity figure, names a technique, gives a wavelength,… Go to post

Worth separating two things that post #8 runs together.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

15 likes in reply to #1 22mo
SR
s.rasmussenTL2 Moderator7 Oct 2024#13
w.novak, post #3: Coming back to the opening post, because the follow-up matters more than the original answer. Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples. Go to post

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

0 likes in reply to #3 22mo
FW
f.wojcikTL2 Moderator15 Oct 2024#14

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

1 like 21mo
J
JFitzgibbonTL2Member22 Oct 2024 · edited#15

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

9 likes 21mo
NN
n.nakamuraTL2 Moderator29 Oct 2024#16
f.wojcik, post #14: Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong. Go to post

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

22 likes in reply to #14 21mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Coming back to: Specificity, linearity, accuracy, precision, range, robustness — with real criteria
On the subject in the title: Specificity, linearity, accuracy, precision, range, robustness — with real criteria Working notes rather than a conclusion. Posting the method first, because I know what the first…
AWHNFJCOT+93 99 1.4k 6d
Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on
Asking directly, because I could not find a straight answer: Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on Working through the identity arithmetic and I would like it…
MSFKJBMRCR+10 14 13k 5mo
An inter-laboratory comparison on one lot, tabulated
On the subject in the title: An inter-laboratory comparison on one lot, tabulated Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a…
ZAMGCRKB 3 17k 10mo
Forced degradation studies: designing one that is informative
Posting this under the heading it deserves: Forced degradation studies: designing one that is informative Everything below is what sits behind that. Working through the identity arithmetic and I would like it…
MPTIEMHKSL+100 106 35k 2d
Robustness: the parameters worth deliberately varying
Robustness: the parameters worth deliberately varying Writing it up because I had to work it out twice and would rather nobody else did. A documentation question rather than an analytical one. I have a…
EFBVPRPOL+53 57 664 8mo

Related topics — sharing the tags LOD & LOQ, precision & repeatability, stability-indicating method

TopicParticipantsRepliesViewsActivity
When to suspect the method rather than the sample
When to suspect the method rather than the sample — that is the question, and I have not found it answered plainly anywhere I have looked. Posting the method first, because I know what the first three replies…
MMCCYMRMBJ+109 115 63k 15mo
Reflectance and photometric devices: precision in practice — a second dataset
Reflectance and photometric devices: precision in practice — a second dataset — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I…
IGCOHSFAFP+75 82 17k 12h
Revisiting: Spike-and-recover accuracy for a peptide assay
Revisiting: Spike-and-recover accuracy for a peptide assay — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it anywhere.…
SCVKEC 2 6.8k 21h
Relative retention time as a better inter-laboratory comparator
On the subject in the title: Relative retention time as a better inter-laboratory comparator Working notes rather than a conclusion. Posting the method first, because I know what the first three replies will…
MCVKSKHVCL+7 11 28k 14mo
Reflectance and photometric devices: precision in practice
On the subject in the title: Reflectance and photometric devices: precision in practice Working notes rather than a conclusion. I would like to understand what this number means before I repeat it anywhere. A…
RMGVB 2 2.2k 10mo