Analytics · Mass spectrometry

Second pass at: Counter-ion content and its effect on measured mass

Solved
Solved by f.fenwick in post #4
Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

Jump to the accepted answer →

SO
sa.okonkwoTL2 Moderator5 Sep 2025#1

Second pass at: Counter-ion content and its effect on measured mass Writing it up because I had to work it out twice and would rather nobody else did.

Working through the identity arithmetic and I would like it checked.

retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4731.3 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

47 likes 11mo
RS
r.scholtenTL2Member13 Nov 2025#2

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

12 likes 8mo
LA
l.aguirreTL2 Moderator1 Jan 2026#3

I read the opening post twice before replying, because I had assumed the opposite.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

1 like 7mo
FF
f.fenwickTL3Regular Solution14 Feb 2026#4

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

7 likes 5mo
OV
o.vogelTL2 Moderator27 Mar 2026#5
f.fenwick, post #4: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

18 likes in reply to #4 4mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Distinguishing a deletion sequence by mass alone — the long version
On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion. I would like to understand what this number means before I repeat it…
TFVKPFIDLF+38 44 60k 11mo
Second pass at: MALDI versus ESI for a quick identity check
Second pass at: MALDI versus ESI for a quick identity check Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it…
MALKO 2 18k 7mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
RMLSMHMINM+121 129 25k 17mo
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo

Related topics — sharing the tags mass error (ppm), counter-ion content, MALDI-TOF

TopicParticipantsRepliesViewsActivity
Deconvolution of a multiply charged envelope
Deconvolution of a multiply charged envelope — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
IDIAJV 2 63 10mo
Endotoxin and why it is a separate question from purity — what changed since
On the subject in the title: Endotoxin and why it is a separate question from purity — what changed since Working notes rather than a conclusion. A documentation question rather than an analytical one. I have…
JSPRMVMDEM+78 91 1k 10h
Coming back to: Deconvolution of a multiply charged envelope
On the subject in the title: Deconvolution of a multiply charged envelope Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
VRCVEL 2 12k 22mo
Second pass at: Where impurities in solid-phase peptide synthesis come from
Second pass at: Where impurities in solid-phase peptide synthesis come from Writing it up because I had to work it out twice and would rather nobody else did. A documentation question rather than an…
GHDBEVLSMP+4 8 7.2k 1mo
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
HEGTMI 2 293 3mo