Validating a method you did not develop — does this still hold?

Solved
Solved by r.bakken in post #4
This follows the opening post rather than contradicting it. Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

Jump to the accepted answer →

AA
a.almeidaTL2 Moderator10 Feb 2025#1

Validating a method you did not develop — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below.

I would like to understand what this number means before I repeat it anywhere.

A Janoshik report on a semaglutide lot gives 96.7% purity. The supplier certificate for the same lot states 98.1%. Both documents name a reversed-phase method; neither states the same gradient.

My question is not "who is right". It is: given that those two figures were produced by different methods, what is the largest difference I should expect from method alone, and at what point does a gap stop being explainable that way?

0 likes 18mo
EF
e.ferreiraTL3Regular11 Feb 2025 · edited#2

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

28 likes 18mo
NR
n.rowntreeTL3Regular11 Feb 2025#3

I read the opening post twice before replying, because I had assumed the opposite.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

9 likes 18mo
RB
r.bakkenTL2 Moderator Solution12 Feb 2025#4
e.ferreira, post #2: Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99. Go to post

This follows the opening post rather than contradicting it.

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

9 likes in reply to #2 17mo
FE
footnote_entryTL3Regular12 Feb 2025#5

On the opening post — agreed on the reasoning, with one qualification.

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

0 likes 17mo
KK
k.kuuselaTL2 Moderator13 Feb 2025#6

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

21 likes 17mo
TN
t.nardoneTL3Regular13 Feb 2025#7

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

5 likes 17mo
NA
n.achebeTL2 Moderator13 Feb 2025#8
footnote_entry, post #5: On the opening post — agreed on the reasoning, with one qualification. Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

Picking up post #5: that is the part I would want checked first.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes in reply to #5 17mo
TD
titration_diaryTL314 Feb 2025#9
HF
h.falkTL2 Moderator14 Feb 2025#10

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

14 likes 17mo
JT
j.teixeiraTL2 Moderator15 Feb 2025#11

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

21 likes 17mo
GI
g.ibarraTL2 Moderator15 Feb 2025#12

On post #8 — agreed on the reasoning, with one qualification.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes 17mo
SC
s.cardosoTL215 Feb 2025#13
BP
b.petrovTL2 Moderator16 Feb 2025#14
e.ferreira, post #2: Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99. Go to post

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

9 likes in reply to #2 17mo
JN
j.nwosuTL2 Moderator16 Feb 2025#15
n.achebe, post #8: Picking up post #5: that is the part I would want checked first. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

post #14 is right about the mechanism and I think understates the practical bit.

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

15 likes in reply to #8 17mo
NS
n.stanescuTL2 Moderator16 Feb 2025#16

Worth separating two things that post #12 runs together.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

30 likes 17mo
NH
n.haddadTL2 Moderator17 Feb 2025 · edited#17

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

1 like 17mo
SK
s.kimaniTL2 Moderator17 Feb 2025#18

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

6 likes 17mo
FI
f.ibarraTL2 Moderator17 Feb 2025#19
footnote_entry, post #5: On the opening post — agreed on the reasoning, with one qualification. Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

post #18 answers the question as asked. The question underneath it is different.

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

1 like in reply to #5 17mo
O
OkaforTL3Regular18 Feb 2025 · edited#20

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

6 likes 17mo
AP
au.pereiraTL2 Moderator18 Feb 2025#21
g.ibarra, post #12: On post #8 — agreed on the reasoning, with one qualification. Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating. Go to post

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

17 likes in reply to #12 17mo
TY
two_year_lineTL3Regular18 Feb 2025#22
n.achebe, post #8: Picking up post #5: that is the part I would want checked first. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

7 likes in reply to #8 17mo
KP
k.pereiraTL2 Moderator19 Feb 2025#23

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

1 like 17mo
MM
maintenance_modeTL3Regular19 Feb 2025 · edited#24

This follows post #21 rather than contradicting it.

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

0 likes 17mo
HB
h.brandtTL2 Moderator19 Feb 2025#25
g.ibarra, post #12: On post #8 — agreed on the reasoning, with one qualification. Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating. Go to post

On post #21 — agreed on the reasoning, with one qualification.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

12 likes in reply to #12 17mo
RV
r.venkatesanTL3Wiki editor20 Feb 2025#26

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

4 likes 17mo
NK
ni.kravchenkoTL2 Moderator20 Feb 2025#27

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes 17mo
IS
isotonic_sheetTL3Regular20 Feb 2025#28

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

0 likes 17mo
RS
r.sobczakTL2 Moderator21 Feb 2025#29

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

32 likes 17mo
R
RodriguesTL3Regular21 Feb 2025#30
footnote_entry, post #5: On the opening post — agreed on the reasoning, with one qualification. Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

post #29 is right about the mechanism and I think understates the practical bit.

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

17 likes in reply to #5 17mo