k.dahlbergTL2 Moderator11 Sep 2024#31
stopper_trace, post #21: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

9 likes in reply to #21 23mo
OB
owen.bradyTL4 Moderator13 Sep 2024#32

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

21 likes 22mo
GR
g.rasmussenTL2 Moderator15 Sep 2024#33

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

0 likes 22mo
AR
a.reyesTL4 Admin16 Sep 2024#34
e.almeida, post #15: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Worth separating two things that post #30 runs together.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

0 likes in reply to #15 22mo
BD
b.demirTL2 Moderator18 Sep 2024#35

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

5 likes 22mo
SB
s.bruunTL2 Moderator20 Sep 2024#36

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

14 likes 22mo
CD
c.delgadoTL2 Moderator21 Sep 2024#37

post #36 answers the question as asked. The question underneath it is different.

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

29 likes 22mo
AL
a.lindholmTL2 Moderator23 Sep 2024 · edited#38
stopper_trace, post #21: System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5). Go to post

On post #34 — agreed on the reasoning, with one qualification.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

0 likes in reply to #21 22mo
RZ
ro.zielinskiTL2 Moderator25 Sep 2024#39

Linearity: the detector response is proportional to compound concentration across the working range. Demonstrated by running standards at multiple concentrations and showing R-squared values typically ≥0.99.

2 likes 22mo
PN
priorauth_notesTL2Regular26 Sep 2024#40

I read post #38 twice before replying, because I had assumed the opposite.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

9 likes 22mo
RE
r.erdoganTL2 Moderator28 Sep 2024#41

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

6 likes 22mo
LG
lc_gradientTL3Analytical chemist29 Sep 2024#42
n.kaufmann, post #24: This follows post #21 rather than contradicting it. Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements. Go to post

Picking up post #39: that is the part I would want checked first.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

1 like in reply to #24 22mo
SO
s.okaforTL2 Moderator1 Oct 2024 · edited#43

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

32 likes 22mo
RA
r.aldana_pharmdTL4Pharmacist3 Oct 2024#44

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

17 likes 22mo
EV
e.vargaTL2 Moderator4 Oct 2024#45

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

3 likes 22mo
BN
bench_notesTL4 Moderator6 Oct 2024#46
m.guerrero, post #22: Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating. Go to post
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes in reply to #22 22mo
KP
k.perrinTL2 Moderator7 Oct 2024#47
ni.kravchenko, post #12: Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples. Go to post

Worth separating two things that post #43 runs together.

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

24 likes in reply to #12 22mo
AB
a.batistaTL2 Moderator9 Oct 2024#48

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

11 likes 22mo
AN
a.nwosuTL2 Moderator10 Oct 2024#49
dexa_twice_yearly, post #29: Worth separating two things that post #25 runs together. Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

0 likes in reply to #29 22mo
DT
d.tammTL212 Oct 2024#50
CN
c.niemelTL3Regular13 Oct 2024#51

post #50 is right about the mechanism and I think understates the practical bit.

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

2 likes 21mo
SF
s.ferreiraTL2 Moderator15 Oct 2024 · edited#52

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

8 likes 21mo
OC
o.cousineauTL317 Oct 2024#53
NV
n.vogelTL2 Moderator18 Oct 2024#54
owen.brady, post #32: Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically. Go to post

I read post #52 twice before replying, because I had assumed the opposite.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

0 likes in reply to #32 21mo
T
TavaresTL1Member20 Oct 2024#55

post #54 answers the question as asked. The question underneath it is different.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

4 likes 21mo
RC
r.chukwuTL2 Moderator21 Oct 2024#56

On post #52 — agreed on the reasoning, with one qualification.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

12 likes 21mo
B
BramleyTL2Member23 Oct 2024#57
p.novak, post #16: Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance. Go to post

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

0 likes in reply to #16 21mo
RM
r.mwangiTL2 Moderator24 Oct 2024#58
dexa_twice_yearly, post #29: Worth separating two things that post #25 runs together. Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

0 likes in reply to #29 21mo
BP
bench_peakTL3Regular26 Oct 2024 · edited#59
lc_gradient, post #42: Picking up post #39: that is the part I would want checked first. Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal). Go to post

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

0 likes in reply to #42 21mo
MA
m.adebayoTL2 Moderator27 Oct 2024#60

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

2 likes 21mo