Why mass spectrometry cannot tell you about an epimer

This is a generated summary. It shows the 9 most-liked posts from a topic of 83, in their original order, with the accepted answer included where one exists. It is a reading aid and it will miss nuance — the full topic is the record.
AA
a.aguirreTL2 Moderator10 Dec 2024 · edited#2

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

27 likes 20mo
GP
g.pemberton_ukTL3Regional · UK25 Dec 2024#11

post #10 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

30 likes 19mo
VK
v.kjaerTL2 Moderator12 Jan 2025#25
p.iyer_pharmd, post #3: I read the opening post twice before replying, because I had assumed the opposite. Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

29 likes in reply to #3 18mo
BK
b.kowalskiTL2 Moderator27 Jan 2025#39

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

31 likes 18mo
ME
me.eriksenTL2 Moderator5 Feb 2025#48

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

27 likes 18mo
ID
integrator_draftTL3Regular8 Feb 2025#51
k.batista, post #14: Coming back to post #12, because the follow-up matters more than the original answer. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

post #50 is right about the mechanism and I think understates the practical bit.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

28 likes in reply to #14 18mo
CM
c.marchettiTL2 Moderator17 Feb 2025#60
stopper_trace, post #17: This follows post #14 rather than contradicting it. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

29 likes in reply to #17 17mo
PF
p.friskTL2 Moderator21 Feb 2025#65

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

31 likes 17mo
L
LJankowiakTL3Regular27 Feb 2025 · edited#71
n.okwuosa, post #4: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

33 likes in reply to #4 17mo

Read the full topic (83 posts)

Suggested topics

TopicParticipantsRepliesViewsActivity
Reading an MS report that only gives you a single number
On the subject in the title: Reading an MS report that only gives you a single number Working notes rather than a conclusion. I would like to understand what this number means before I repeat it anywhere. A…
BDSBCDAMBR+100 113 48k 13mo
Mass error in ppm: the arithmetic and the caveat — a second dataset
On the subject in the title: Mass error in ppm: the arithmetic and the caveat — a second dataset Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked.…
HKGVDWONW+20 24 1.7k 1d
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold?
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold? — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity…
PODBTD 2 4.6k 7h
Adducts: sodium, potassium, and the peak you did not expect — does this still hold?
Asking directly, because I could not find a straight answer: Adducts: sodium, potassium, and the peak you did not expect — does this still hold? Posting the method first, because I know what the first three…
KBZYLVCETV+113 119 1.7k 3mo
Adducts: sodium, potassium, and the peak you did not expect
Adducts: sodium, potassium, and the peak you did not expect Writing it up because I had to work it out twice and would rather nobody else did. I would like to understand what this number means before I repeat…
HMDHNCJWSV+70 77 3.5k 15mo

Related topics — sharing the tags LC-MS, mass error (ppm), high-resolution MS

TopicParticipantsRepliesViewsActivity
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
HEGTMI 2 293 3mo
Second pass at: MALDI versus ESI for a quick identity check
Second pass at: MALDI versus ESI for a quick identity check Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it…
MALKO 2 18k 7mo
Distinguishing a deletion sequence by mass alone — the long version
On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion. I would like to understand what this number means before I repeat it…
TFVKPFIDLF+38 44 60k 11mo
Reading an MS report that only gives you a single number — the long version
Reading an MS report that only gives you a single number — the long version — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I…
TTRAK 2 47k 18mo
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Working through the identity arithmetic and I would like it checked.…
CRGISCBPEC+117 123 6.5k 8mo