k.perrinTL2 Moderator18 Jan 2025 · edited#31

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 18mo
BN
bench_notesTL4 Moderator19 Jan 2025#32
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

3 likes 18mo
AN
a.nwosuTL2 Moderator21 Jan 2025#33

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

10 likes 18mo
AB
a.batistaTL2 Moderator22 Jan 2025#34
h.jansen, post #27: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Worth separating two things that post #30 runs together.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

22 likes in reply to #27 18mo
RV
r.villalobosTL2 Moderator23 Jan 2025#35

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 18mo
RD
r.danquahTL224 Jan 2025#36
MD
m.duarteTL2 Moderator25 Jan 2025#37
v.kjaer, post #25: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

post #36 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

6 likes in reply to #25 18mo
OV
o.vogelTL2 Moderator26 Jan 2025#38
a.batista, post #34: Worth separating two things that post #30 runs together. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

On post #34 — agreed on the reasoning, with one qualification.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

16 likes in reply to #34 18mo
BK
b.kowalskiTL2 Moderator27 Jan 2025#39

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

31 likes 18mo
EF
e.ferreiraTL3Regular28 Jan 2025#40

I read post #38 twice before replying, because I had assumed the opposite.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 18mo
NB
n.brobergTL2 Moderator29 Jan 2025#41
h.iyer, post #6: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

19 likes in reply to #6 18mo
SK
s.karlsen_rphTL3Pharmacist30 Jan 2025#42

Picking up post #39: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

8 likes 18mo
MP
m.perrinTL2 Moderator31 Jan 2025#43

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 18mo
DO
dr_okonkwoTL4 Moderator1 Feb 2025#44
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 18mo
FS
f.sjobergTL2 Moderator2 Feb 2025 · edited#45
a.aguirre, post #2: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

13 likes in reply to #2 18mo
CC
c.cardosoTL2 Moderator3 Feb 2025#46
m.duarte, post #37: post #36 answers the question as asked. The question underneath it is different. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

4 likes in reply to #37 18mo
ML
m.lehtinenTL2 Moderator4 Feb 2025#47

Worth separating two things that post #43 runs together.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes 18mo
ME
me.eriksenTL2 Moderator5 Feb 2025#48

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

27 likes 18mo
CC
c.correiaTL2 Moderator6 Feb 2025#49

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 18mo
RR
r.restrepoTL2 Moderator7 Feb 2025#50

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

18 likes 18mo
ID
integrator_draftTL3Regular8 Feb 2025#51
k.batista, post #14: Coming back to post #12, because the follow-up matters more than the original answer. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

post #50 is right about the mechanism and I think understates the practical bit.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

28 likes in reply to #14 18mo
YR
y.ramosTL2 Moderator9 Feb 2025 · edited#52

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

0 likes 18mo
VK
v.klausenTL3Regular10 Feb 2025#53

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

5 likes 18mo
CH
ca.haddadTL2 Moderator11 Feb 2025#54

I read post #52 twice before replying, because I had assumed the opposite.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

13 likes 18mo
AS
a.stephanopoulosTL3Regular12 Feb 2025#55

post #54 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 17mo
FI
f.ibarraTL2 Moderator13 Feb 2025#56

On post #52 — agreed on the reasoning, with one qualification.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 17mo
O
OkaforTL3Regular14 Feb 2025#57

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

8 likes 17mo
NS
n.szaboTL2 Moderator15 Feb 2025#58
me.eriksen, post #48: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

19 likes in reply to #48 17mo
D
DOdendaalTL3Regular16 Feb 2025#59
n.okwuosa, post #4: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

14 likes in reply to #4 17mo
CM
c.marchettiTL2 Moderator17 Feb 2025#60
stopper_trace, post #17: This follows post #14 rather than contradicting it. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

29 likes in reply to #17 17mo