orbitrap_olaTL3Mass spectrometrist10 Jun 2025#61

post #60 answers the question as asked. The question underneath it is different.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

2 likes 14mo
IA
i.almeidaTL2 Moderator11 Jun 2025 · edited#62

On post #58 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

8 likes 14mo
DS
dr_seongTL3Physician13 Jun 2025#63
s.girard, post #55: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

19 likes in reply to #55 13mo
CV
c.vasquezTL214 Jun 2025#64
CL
customs_ledgerTL3Regular15 Jun 2025#65

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

4 likes 13mo
FW
f.weissTL2 Moderator16 Jun 2025#66

Worth separating two things that post #62 runs together.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

12 likes 13mo
WN
w.novakTL3Regular17 Jun 2025#67
m.nascimento, post #51: Worth separating two things that post #47 runs together. Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing. The correction was fair and I had been repeating something I had not checked carefully enough. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

26 likes in reply to #51 13mo
NK
n.kravchenkoTL2 Moderator18 Jun 2025#68
w.novak, post #67: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes in reply to #67 13mo
ME
me.eriksenTL219 Jun 2025#69
ML
m.lehtinenTL2 Moderator21 Jun 2025#70

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

18 likes 13mo
N
NicolaidesTL3Regular22 Jun 2025#71

I read post #69 twice before replying, because I had assumed the opposite.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

7 likes 13mo
WV
w.verhoevenTL2 Moderator23 Jun 2025#72
Okafor, post #6: Resolution: "high resolution" commonly means Go to post

This follows post #69 rather than contradicting it.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

1 like in reply to #6 13mo
SG
s.grigorescuTL2Member24 Jun 2025#73

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes 13mo
RW
r.weissTL2 Moderator25 Jun 2025#74

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

24 likes 13mo
D
DKwiatkowskiTL3Regular26 Jun 2025#75

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

11 likes 13mo
DA
d.achebeTL2 Moderator27 Jun 2025#76

Picking up post #73: that is the part I would want checked first.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

3 likes 13mo
N
NorringtonTL3Regular28 Jun 2025 · edited#77
f.danquah, post #57: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes in reply to #57 13mo
GT
g.tammTL2 Moderator29 Jun 2025#78

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

32 likes 13mo
WP
weekly_pinTL2Regular1 Jul 2025#79
b.osei, post #40: Picking up post #37: that is the part I would want checked first. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

16 likes in reply to #40 13mo
SA
s.adebayoTL2 Moderator2 Jul 2025#80
m.yildiz, post #23: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

6 likes in reply to #23 13mo
KB
k.brandl_deTL3Translator · DE3 Jul 2025#81

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

6 likes 13mo
VB
v.bruunTL2 Moderator4 Jul 2025#82

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

16 likes 13mo
AK
a.kowalczykTL2Regular5 Jul 2025#83

post #82 answers the question as asked. The question underneath it is different.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes 13mo
MA
m.almeidaTL2 Moderator6 Jul 2025 · edited#84
t.nardone, post #18: This follows post #15 rather than contradicting it. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

On post #80 — agreed on the reasoning, with one qualification.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

1 like in reply to #18 13mo
G
GEldridgeTL3Regular7 Jul 2025#85
mi.amankwah, post #46: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

3 likes in reply to #46 13mo
HJ
h.jansenTL2 Moderator8 Jul 2025#86

I read post #84 twice before replying, because I had assumed the opposite.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

11 likes 13mo
DB
d.bramleyTL3Regular9 Jul 2025#87

post #86 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

32 likes 13mo
VK
v.kjaerTL2 Moderator10 Jul 2025#88

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 13mo
R
RidgewayTL3Regular11 Jul 2025#89

Picking up post #86: that is the part I would want checked first.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

1 like 13mo
ZA
z.adeyemiTL2 Moderator12 Jul 2025#90

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

7 likes 13mo