Counter-ion content and its effect on measured mass posts 31–60

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

AM
a.mwangiTL2 Moderator4 May 2025 · edited#31

Coming back to post #29, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

15 likes 15mo
CD
cannula_driftTL3Regular5 May 2025#32

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

6 likes 15mo
SV
sa.vogelTL2 Moderator7 May 2025#33
j.marchetti, post #27: Picking up post #24: that is the part I would want checked first. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

1 like in reply to #27 15mo
BR
buffer_reviewTL3Regular8 May 2025#34
HHidalgo, post #20: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

post #33 answers the question as asked. The question underneath it is different.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes in reply to #20 15mo
HB
h.bhattacharyaTL2 Moderator9 May 2025#35

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

10 likes 15mo
C
CSagredoTL3Regular11 May 2025#36

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

3 likes 15mo
RN
r.novakTL212 May 2025#37
OA
o.abrahamsenTL3Regular13 May 2025#38

post #37 is right about the mechanism and I think understates the practical bit.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

31 likes 15mo
AR
a.reyesTL4 Admin15 May 2025#39
il.dumitru, post #3: Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

6 likes in reply to #3 14mo
BO
b.oseiTL2 Moderator16 May 2025 · edited#40

Picking up post #37: that is the part I would want checked first.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

1 like 14mo
BR
buffer_reviewTL3Regular17 May 2025#41

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

9 likes 14mo
AN
a.norgaardTL2 Moderator18 May 2025#42

I read post #40 twice before replying, because I had assumed the opposite.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

21 likes 14mo
L
LJankowiakTL3Regular20 May 2025#43
a.reyes, post #39: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

post #42 is right about the mechanism and I think understates the practical bit.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes in reply to #39 14mo
AC
a.cardosoTL2 Moderator21 May 2025 · edited#44

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

2 likes 14mo
AD
ambient_draftTL3Regular22 May 2025#45

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

14 likes 14mo
MA
mi.amankwahTL2 Moderator23 May 2025#46
lyophil_margin, post #22: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

29 likes in reply to #22 14mo
TS
t.steenkampTL2Member25 May 2025#47
o.abrahamsen, post #38: post #37 is right about the mechanism and I think understates the practical bit. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of… Go to post

post #46 answers the question as asked. The question underneath it is different.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #38 14mo
AK
ak.kravchenkoTL2 Moderator26 May 2025#48

On post #44 — agreed on the reasoning, with one qualification.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

5 likes 14mo
P
PWendelboeTL1Member27 May 2025#49

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

3 likes 14mo
AW
ai.wikstromTL2 Moderator28 May 2025#50
st.diallo, post #19: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

10 likes in reply to #19 14mo
MN
m.nascimentoTL2 Moderator29 May 2025#51

Worth separating two things that post #47 runs together.

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

5 likes 14mo
AF
a.finnegan_rdTL2Dietitian31 May 2025#52

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 14mo
PO
p.ostergaardTL2 Moderator1 Jun 2025#53
CSagredo, post #36: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

28 likes in reply to #36 14mo
CL
coldchain_liuTL3Regular2 Jun 2025#54

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

14 likes 14mo
SG
s.girardTL2 Moderator3 Jun 2025#55

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

2 likes 14mo
LE
logbook_erinTL3Regular4 Jun 2025#56

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes 14mo
FD
f.danquahTL2 Moderator6 Jun 2025#57
a.cardoso, post #44: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

21 likes in reply to #44 14mo
CO
c.okaforTL3Regular7 Jun 2025 · edited#58
cannula_drift, post #32: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Picking up post #55: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

9 likes in reply to #32 14mo
ND
n.duarteTL2 Moderator8 Jun 2025#59

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

13 likes 14mo
CC
crossref_checkTL3Wiki editor9 Jun 2025#60

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

4 likes 14mo