Mass error in ppm: the arithmetic and the caveat

Solved
Solved by b.vestergaard in post #5
On the opening post — agreed on the reasoning, with one qualification. Resolution: "high resolution" commonly means

Jump to the accepted answer →

KC
k.chukwuTL2 Moderator11 Feb 2026#1

Mass error in ppm: the arithmetic and the caveat Writing it up because I had to work it out twice and would rather nobody else did.

Working through the identity arithmetic and I would like it checked.

retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4731.3 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

17 likes 5mo
CI
c.inglethorpeTL3Regular11 Feb 2026#2

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

2 likes 5mo
AM
a.mwangiTL2 Moderator11 Feb 2026#3

I read the opening post twice before replying, because I had assumed the opposite.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes 5mo
CD
cannula_driftTL3Regular11 Feb 2026#4

This follows post #3 rather than contradicting it.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

20 likes 5mo
BV
b.vestergaardTL2 Moderator Solution11 Feb 2026#5
k.chukwu, post #1: Mass error in ppm: the arithmetic and the caveat Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked. retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series… Go to post

On the opening post — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

12 likes in reply to #1 5mo
B
BGiordanoTL2Member11 Feb 2026#6
b.vestergaard, post #5: On the opening post — agreed on the reasoning, with one qualification. Resolution: "high resolution" commonly means Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes in reply to #5 5mo
AL
a.lindholmTL2 Moderator11 Feb 2026#7

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

29 likes 5mo
CD
c.delgadoTL2 Moderator11 Feb 2026 · edited#8

Picking up post #5: that is the part I would want checked first.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

14 likes 5mo
PL
p.lindqvistTL2 Moderator11 Feb 2026#9
a.lindholm, post #7: Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

2 likes in reply to #7 5mo
CE
crossover_entryTL3Regular11 Feb 2026#10

post #9 is right about the mechanism and I think understates the practical bit.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 5mo
KB
k.batistaTL2 Moderator12 Feb 2026#11

post #10 answers the question as asked. The question underneath it is different.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

1 like 5mo
MM
methods_marginTL312 Feb 2026#12
MG
m.guerreroTL2 Moderator12 Feb 2026#13

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

23 likes 5mo
ST
stopper_traceTL2Member12 Feb 2026#14
m.guerrero, post #13: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes in reply to #13 5mo
NH
n.hartmannTL2 Moderator12 Feb 2026#15
BGiordano, post #6: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

post #14 is right about the mechanism and I think understates the practical bit.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #6 5mo
VS
vial_slopeTL3Regular12 Feb 2026 · edited#16

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

3 likes 5mo
MA
m.amankwahTL2 Moderator12 Feb 2026#17

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

16 likes 5mo
N
NLoughranTL3Regular12 Feb 2026#18

I read post #16 twice before replying, because I had assumed the opposite.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

32 likes 5mo
RV
r.vukovicTL2 Moderator12 Feb 2026#19
c.inglethorpe, post #2: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes in reply to #2 5mo
RM
r.mcalisterTL3Regular12 Feb 2026#20

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

1 like 5mo
VK
v.klausenTL3Regular12 Feb 2026 · edited#21

Worth separating two things that post #17 runs together.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 5mo
EC
e.coelhoTL2 Moderator12 Feb 2026#22

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes 5mo
G
GDashwoodTL3Regular12 Feb 2026#23
r.vukovic, post #19: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

19 likes in reply to #19 5mo
CH
ca.haddadTL2 Moderator12 Feb 2026#24

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

8 likes 5mo
GI
g.ibarraTL2 Moderator12 Feb 2026#25

On post #21 — agreed on the reasoning, with one qualification.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 5mo
JT
j.teixeiraTL2 Moderator12 Feb 2026#26
a.mwangi, post #3: I read the opening post twice before replying, because I had assumed the opposite. Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

post #25 answers the question as asked. The question underneath it is different.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

27 likes in reply to #3 5mo
BP
b.petrovTL2 Moderator12 Feb 2026#27

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

13 likes 5mo
SC
s.cardosoTL2 Moderator12 Feb 2026#28

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

4 likes 5mo
NS
n.stanescuTL2 Moderator12 Feb 2026#29

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

4 likes 5mo
JN
j.nwosuTL2 Moderator12 Feb 2026#30

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.