Analytics · Mass spectrometry · continued

Mass error in ppm: the arithmetic and the caveat posts 61–90

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

ES
e.silvaTL2 Moderator14 Feb 2026#61
a.lindholm, post #7: Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes in reply to #7 5mo
LF
l.ferreiraTL2 Moderator14 Feb 2026#62
j.fonseca, post #55: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes in reply to #55 5mo
VS
v.salgadoTL2 Moderator14 Feb 2026#63

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

17 likes 5mo
AA
an.adeyemiTL2 Moderator14 Feb 2026#64

Picking up post #61: that is the part I would want checked first.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

7 likes 5mo
VK
v.klausenTL3Regular14 Feb 2026#65

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

1 like 5mo
HF
h.fonsecaTL2 Moderator14 Feb 2026#66
v.klausen, post #65: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

0 likes in reply to #65 5mo
HK
h.koodziejTL2Member14 Feb 2026#67

I read post #65 twice before replying, because I had assumed the opposite.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

24 likes 5mo
SS
s.solbergTL2 Moderator14 Feb 2026#68

This follows post #65 rather than contradicting it.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

11 likes 5mo
VM
v.milanoviTL3Regular14 Feb 2026 · edited#69
l.ferreira, post #62: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

On post #65 — agreed on the reasoning, with one qualification.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes in reply to #62 5mo
FP
f.petrovTL214 Feb 2026#70
SS
s.stavrianosTL2Member14 Feb 2026#71

Picking up post #68: that is the part I would want checked first.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

2 likes 5mo
GD
g.danquahTL2 Moderator14 Feb 2026#72

Coming back to post #70, because the follow-up matters more than the original answer.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

8 likes 5mo
GC
glossary_checkTL2Member14 Feb 2026#73

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

19 likes 5mo
SP
s.perrinTL2 Moderator14 Feb 2026#74
c.inglethorpe, post #2: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes in reply to #2 5mo
TW
t.waldenstrmTL2Member14 Feb 2026 · edited#75
k.batista, post #11: post #10 answers the question as asked. The question underneath it is different. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of… Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

4 likes in reply to #11 5mo
SR
s.radichTL2 Moderator14 Feb 2026#76

I read post #74 twice before replying, because I had assumed the opposite.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

12 likes 5mo
B
BuchholzTL2Member14 Feb 2026#77

post #76 is right about the mechanism and I think understates the practical bit.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

26 likes 5mo
EK
ew.kuuselaTL2 Moderator14 Feb 2026#78
s.stavrianos, post #71: Picking up post #68: that is the part I would want checked first. Resolution: "high resolution" commonly means Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes in reply to #71 5mo
AL
aliquot_lineTL3Regular14 Feb 2026#79
a.mwangi, post #3: I read the opening post twice before replying, because I had assumed the opposite. Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #3 5mo
AK
a.krastevTL2 Moderator14 Feb 2026#80

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

2 likes 5mo
AA
an.adeyemiTL214 Feb 2026#81
ES
e.silvaTL2 Moderator14 Feb 2026#82

This follows post #79 rather than contradicting it.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

0 likes 5mo
LF
l.ferreiraTL2 Moderator14 Feb 2026#83

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

15 likes 5mo
HK
h.koodziejTL2Member14 Feb 2026#84
e.coelho, post #22: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

6 likes in reply to #22 5mo
JC
j.castellanosTL2 Moderator14 Feb 2026#85

Coming back to post #83, because the follow-up matters more than the original answer.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

2 likes 5mo
JS
j.sorensenTL2 Moderator15 Feb 2026#86

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 5mo
CR
c.rasmussenTL2 Moderator15 Feb 2026#87

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

21 likes 5mo
EL
e.lokkenTL2 Moderator15 Feb 2026#88
r.erdogan, post #33: post #32 is right about the mechanism and I think understates the practical bit. For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

post #87 answers the question as asked. The question underneath it is different.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

9 likes in reply to #33 5mo
C
chromatogramTL4Analytical chemist15 Feb 2026#89
m.guerrero, post #13: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

5 likes in reply to #13 5mo
AW
a.wikstromTL215 Feb 2026#90