BK
b.kowalskiTL2 Moderator12 Feb 2026#31
c.inglethorpe, post #2: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

2 likes in reply to #2 5mo
EF
e.ferreiraTL3Regular12 Feb 2026#32
p.lindqvist, post #9: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

I read post #30 twice before replying, because I had assumed the opposite.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

9 likes in reply to #9 5mo
RE
r.erdoganTL2 Moderator13 Feb 2026#33

post #32 is right about the mechanism and I think understates the practical bit.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

20 likes 5mo
DH
dietitian_hollisTL3Dietitian13 Feb 2026#34

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes 5mo
SM
s.mbekiTL2 Moderator13 Feb 2026 · edited#35
BGiordano, post #6: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Picking up post #32: that is the part I would want checked first.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes in reply to #6 5mo
NA
n.abernathyTL3Analytical chemist13 Feb 2026#36

Coming back to post #34, because the follow-up matters more than the original answer.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

5 likes 5mo
AV
a.vukovicTL2 Moderator13 Feb 2026#37

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

14 likes 5mo
RA
r.aldana_pharmdTL4Pharmacist13 Feb 2026#38

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

29 likes 5mo
KP
k.perrinTL2 Moderator13 Feb 2026#39

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

27 likes 5mo
BN
bench_notesTL4 Moderator13 Feb 2026#40
r.vukovic, post #19: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

19 likes in reply to #19 5mo
SS
steady_stateTL3Regular13 Feb 2026#41

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

5 likes 5mo
YA
y.asanteTL2 Moderator13 Feb 2026#42
bench_notes, post #40: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes in reply to #40 5mo
JW
journalclub_wrenTL3Regular13 Feb 2026#43

On post #39 — agreed on the reasoning, with one qualification.

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

27 likes 5mo
EH
e.halonenTL2 Moderator13 Feb 2026#44

post #43 answers the question as asked. The question underneath it is different.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

13 likes 5mo
DH
dietitian_hollisTL3Dietitian13 Feb 2026 · edited#45

I read post #43 twice before replying, because I had assumed the opposite.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

2 likes 5mo
HA
h.agyemanTL2 Moderator13 Feb 2026#46
k.perrin, post #39: For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes in reply to #39 5mo
NA
n.abernathyTL3Analytical chemist13 Feb 2026#47
n.stanescu, post #29: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

20 likes in reply to #29 5mo
NK
n.kuuselaTL2 Moderator13 Feb 2026#48

post #47 is right about the mechanism and I think understates the practical bit.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

9 likes 5mo
B
BirkelandTL3Regular13 Feb 2026#49

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 5mo
VR
v.rautioTL2 Moderator13 Feb 2026 · edited#50

Picking up post #47: that is the part I would want checked first.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 5mo
IA
i.almeidaTL2 Moderator13 Feb 2026#51

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

0 likes 5mo
DS
dr_seongTL3Physician13 Feb 2026#52

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

1 like 5mo
RF
ro.friskTL2 Moderator13 Feb 2026#53

This follows post #50 rather than contradicting it.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

12 likes 5mo
CL
customs_ledgerTL3Regular13 Feb 2026#54
dr_seong, post #52: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

25 likes in reply to #52 5mo
JF
j.fonsecaTL213 Feb 2026#55
PW
PharmNotes_WhitfieldTL4Pharmacist13 Feb 2026#56

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

4 likes 5mo
NB
n.brobergTL2 Moderator13 Feb 2026#57
a.vukovic, post #37: Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Picking up post #54: that is the part I would want checked first.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

17 likes in reply to #37 5mo
OO
orbitrap_olaTL3Mass spectrometrist14 Feb 2026#58
v.rautio, post #50: Picking up post #47: that is the part I would want checked first. Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up. Go to post

Coming back to post #56, because the follow-up matters more than the original answer.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

0 likes in reply to #50 5mo
HE
h.espinozaTL2 Moderator14 Feb 2026#59

post #58 is right about the mechanism and I think understates the practical bit.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

1 like 5mo
YM
y.mensahTL3Wiki editor14 Feb 2026#60

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

7 likes 5mo