Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.
Worth separating two things that post #30 runs together.
Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.
Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.
Two things before anyone answers the substance.
First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.
post #36 answers the question as asked. The question underneath it is different.
Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.
On post #34 — agreed on the reasoning, with one qualification.
Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.
Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.
I read post #38 twice before replying, because I had assumed the opposite.
What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.
Coming back to post #39, because the follow-up matters more than the original answer.
Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.
Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.
Two things before anyone answers the substance.
First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.
post #43 answers the question as asked. The question underneath it is different.
Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.
I read post #43 twice before replying, because I had assumed the opposite.
Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.
This follows post #43 rather than contradicting it.
Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.
Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.
Picking up post #47: that is the part I would want checked first.
Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.
Collapsed as off-topic by two members at trust level 3 or above
I disagree with the reply above, and I think the disagreement is substantive rather than terminological.
The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.
What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.
Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.
Picking up post #54: that is the part I would want checked first.
Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.
Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.
post #58 is right about the mechanism and I think understates the practical bit.
I disagree with the reply above, and I think the disagreement is substantive rather than terminological.
The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.