Adducts: sodium, potassium, and the peak you did not expect posts 31–60

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

BJ
b.jankowiakTL3Regular10 Mar 2025#31

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

18 likes 17mo
PD
p.dialloTL2 Moderator11 Mar 2025#32
r.scholten, post #19: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes in reply to #19 17mo
YM
y.mensahTL3Wiki editor13 Mar 2025#33

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 17mo
JF
j.falkTL2 Moderator14 Mar 2025#34

Worth separating two things that post #30 runs together.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

4 likes 16mo
ST
slow_titratorTL2Regular15 Mar 2025#35
Ziegler, post #17: Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

12 likes in reply to #17 16mo
HE
h.espinozaTL2 Moderator16 Mar 2025#36
j.delacroix, post #25: Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

26 likes in reply to #25 16mo
WN
w.novakTL3Regular17 Mar 2025#37

post #36 answers the question as asked. The question underneath it is different.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes 16mo
TK
t.karlsenTL2 Moderator18 Mar 2025 · edited#38

On post #34 — agreed on the reasoning, with one qualification.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

2 likes 16mo
ED
e.dalgleishTL3Regular19 Mar 2025#39
steady_state, post #6: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

8 likes in reply to #6 16mo
MB
ma.balogunTL2 Moderator20 Mar 2025#40

I read post #38 twice before replying, because I had assumed the opposite.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

19 likes 16mo
PK
p.krastevTL2 Moderator21 Mar 2025#41

Coming back to post #39, because the follow-up matters more than the original answer.

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

14 likes 16mo
RV
r.venkatesanTL3Wiki editor22 Mar 2025#42

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

5 likes 16mo
KP
k.pereiraTL2 Moderator23 Mar 2025 · edited#43
dietitian_hollis, post #2: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes in reply to #2 16mo
MM
maintenance_modeTL3Regular24 Mar 2025#44
h.espinoza, post #36: Two things before anyone answers the substance. First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound. Go to post

post #43 answers the question as asked. The question underneath it is different.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

29 likes in reply to #36 16mo
AP
au.pereiraTL2 Moderator25 Mar 2025#45

I read post #43 twice before replying, because I had assumed the opposite.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

9 likes 16mo
LE
logbook_erinTL3Regular26 Mar 2025#46

This follows post #43 rather than contradicting it.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

2 likes 16mo
AI
a.ilungaTL2 Moderator27 Mar 2025#47

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 16mo
CL
coldchain_liuTL3Regular28 Mar 2025#48
j.falk, post #34: Worth separating two things that post #30 runs together. Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

21 likes in reply to #34 16mo
JI
j.iyerTL2 Moderator29 Mar 2025#49

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

28 likes 16mo
PM
physio_marchettiTL2Physiotherapist30 Mar 2025#50
n.cabrera, post #3: I read the opening post twice before replying, because I had assumed the opposite. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Picking up post #47: that is the part I would want checked first.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

14 likes in reply to #3 16mo
MB
m.brobergTL2 Moderator31 Mar 2025#51

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

22 likes 16mo
SL
s.leclercTL4 Moderator1 Apr 2025#52

Worth separating two things that post #48 runs together.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes 16mo
AW
a.wikstromTL21 Apr 2025#53
C
chromatogramTL4Analytical chemist2 Apr 2025#54
s.leclerc, post #52: Worth separating two things that post #48 runs together. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

10 likes in reply to #52 16mo
EL
e.lokkenTL2 Moderator3 Apr 2025 · edited#55

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

30 likes 16mo
CR
c.rasmussenTL2 Moderator4 Apr 2025#56

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 16mo
JN
j.nascimentoTL2 Moderator5 Apr 2025#57
w.novak, post #37: post #36 answers the question as asked. The question underneath it is different. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very… Go to post

Picking up post #54: that is the part I would want checked first.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

5 likes in reply to #37 16mo
CB
c.bakkerTL2 Moderator6 Apr 2025#58
physio_marchetti, post #50: Picking up post #47: that is the part I would want checked first. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

15 likes in reply to #50 16mo
K
KTurkingtonTL3Regular7 Apr 2025#59
steady_state, post #6: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

post #58 is right about the mechanism and I think understates the practical bit.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

10 likes in reply to #6 16mo
TD
t.demirTL2 Moderator8 Apr 2025#60

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

23 likes 16mo