CN
cannula_notesTL2Member17 Mar 2025#31

Coming back to post #29, because the follow-up matters more than the original answer.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

5 likes 16mo
AW
am.wikstromTL2 Moderator19 Mar 2025 · edited#32

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes 16mo
I
IbrahimoviTL2Member20 Mar 2025#33
sa.vogel, post #6: Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes in reply to #6 16mo
ZN
z.nakamuraTL2 Moderator21 Mar 2025#34
cannula_notes, post #31: Coming back to post #29, because the follow-up matters more than the original answer. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

post #33 answers the question as asked. The question underneath it is different.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

20 likes in reply to #31 16mo
D
DSakamotoTL3Regular23 Mar 2025#35

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

2 likes 16mo
CK
c.kuuselaTL2 Moderator24 Mar 2025#36

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 16mo
TF
taper_fileTL3Regular25 Mar 2025#37
j.habermann, post #22: Worth separating two things that post #18 runs together. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Worth separating two things that post #33 runs together.

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

28 likes in reply to #22 16mo
HK
h.krastevTL2 Moderator27 Mar 2025#38

post #37 is right about the mechanism and I think understates the practical bit.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

14 likes 16mo
PS
p.silvaTL2 Moderator28 Mar 2025#39

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

0 likes 16mo
MM
m.malinowskiTL2 Moderator29 Mar 2025#40

Picking up post #37: that is the part I would want checked first.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

0 likes 16mo
AM
a.molnarTL2 Moderator31 Mar 2025#41

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

2 likes 16mo
BT
baseline_tableTL2Member1 Apr 2025#42
h.brandt, post #11: On post #7 — agreed on the reasoning, with one qualification. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

8 likes in reply to #11 16mo
AV
a.villalobosTL2 Moderator2 Apr 2025#43

post #42 is right about the mechanism and I think understates the practical bit.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

27 likes 16mo
D
DSakamotoTL33 Apr 2025#44
JL
j.lokkenTL2 Moderator5 Apr 2025#45
a.molnar, post #41: Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

4 likes in reply to #41 16mo
TF
taper_fileTL3Regular6 Apr 2025#46
LJankowiak, post #7: post #6 is right about the mechanism and I think understates the practical bit. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Having read the exchange above, I think I was wrong earlier in this topic and I want to say so plainly rather than quietly editing.

The correction was fair and I had been repeating something I had not checked carefully enough.

13 likes in reply to #7 16mo
DA
d.achebeTL2 Moderator7 Apr 2025#47

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

0 likes 16mo
D
DKwiatkowskiTL3Regular8 Apr 2025 · edited#48

On post #44 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

0 likes 16mo
FP
f.piresTL2 Moderator10 Apr 2025#49
e.mikkelsen, post #24: Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak. Go to post

This follows post #46 rather than contradicting it.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes in reply to #24 16mo
GD
glossary_deskTL3Regular11 Apr 2025#50

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

2 likes 16mo
AP
au.pereiraTL2 Moderator12 Apr 2025#51

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

24 likes 16mo
LE
logbook_erinTL313 Apr 2025#52
FD
f.danquahTL2 Moderator15 Apr 2025#53

I read post #51 twice before replying, because I had assumed the opposite.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

1 like 15mo
MM
maintenance_modeTL3Regular16 Apr 2025#54
a.villalobos, post #43: post #42 is right about the mechanism and I think understates the practical bit. I disagree with the reply above, and I think the disagreement is substantive rather than terminological. The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this,… Go to post

This follows post #51 rather than contradicting it.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes in reply to #43 15mo
PK
p.krastevTL2 Moderator17 Apr 2025#55

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

17 likes 15mo
RV
r.venkatesanTL3Wiki editor18 Apr 2025#56

post #55 answers the question as asked. The question underneath it is different.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

7 likes 15mo
NK
ni.kravchenkoTL2 Moderator19 Apr 2025#57

Coming back to post #55, because the follow-up matters more than the original answer.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 15mo
IS
isotonic_sheetTL3Regular21 Apr 2025#58
p.trevino, post #13: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes in reply to #13 15mo
AV
a.vermeulenTL2 Moderator22 Apr 2025#59
r.oyelaran, post #21: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Worth separating two things that post #55 runs together.

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #21 15mo
RJ
r.jhannsdttirTL3Regular23 Apr 2025 · edited#60

post #59 is right about the mechanism and I think understates the practical bit.

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

23 likes 15mo