Reading an MS report that only gives you a single number posts 91–114

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

ID
i.dumitruTL2 Moderator28 May 2025#91

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

19 likes 14mo
R
RidgewayTL3Regular29 May 2025 · edited#92

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

8 likes 14mo
IG
in.guerreroTL2 Moderator30 May 2025#93

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

0 likes 14mo
EL
endpoint_lineTL3Regular31 May 2025#94
logbook_erin, post #52: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Picking up post #91: that is the part I would want checked first.

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes in reply to #52 14mo
AC
a.cabreraTL2 Moderator1 Jun 2025#95

Worth separating two things that post #91 runs together.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

26 likes 14mo
CP
citation_peakTL3Regular2 Jun 2025#96

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

13 likes 14mo
IG
i.grimaldiTL2 Moderator3 Jun 2025#97
t.wojcik, post #65: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Practical note that does not fit anywhere else. Whatever you conclude from this topic, write down what you did and when. The single most useful thing in your own records is not any individual result; it is that they are dated and consecutive.

2 likes in reply to #65 14mo
EF
erratum_fileTL3Regular4 Jun 2025#98
DSakamoto, post #44: Worth separating two things that post #40 runs together. Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Go to post

Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.

0 likes in reply to #44 14mo
MD
m.dumitruTL2 Moderator5 Jun 2025#99
f.danquah, post #53: I read post #51 twice before replying, because I had assumed the opposite. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this… Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

9 likes in reply to #53 14mo
DW
diluent_watchTL2Member6 Jun 2025#100

post #99 answers the question as asked. The question underneath it is different.

I disagree with the reply above, and I think the disagreement is substantive rather than terminological.

The distinction being drawn does not survive when you look at the published data for this specific question. I would be glad to be shown wrong on this, because the version I am arguing against is more convenient.

2 likes 14mo
HK
h.koodziejTL2Member7 Jun 2025#101

Coming back to post #99, because the follow-up matters more than the original answer.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

0 likes 14mo
EM
e.mwangiTL2 Moderator8 Jun 2025#102
dr_seong, post #73: Worth separating two things that post #69 runs together. Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. Go to post

Picking up post #99: that is the part I would want checked first.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

28 likes in reply to #73 14mo
TI
trough_indexTL3Regular9 Jun 2025#103

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

9 likes 14mo
HF
h.fonsecaTL2 Moderator10 Jun 2025#104

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

2 likes 14mo
NB
n.bridgewaterTL2Member11 Jun 2025#105

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 14mo
NL
ne.laurentTL2 Moderator13 Jun 2025#106
i.guerrero, post #87: post #86 is right about the mechanism and I think understates the practical bit. For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

This follows post #103 rather than contradicting it.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

21 likes in reply to #87 13mo
VM
v.milanoviTL3Regular14 Jun 2025#107

Worth separating two things that post #103 runs together.

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

5 likes 13mo
SL
s.lundgrenTL2 Moderator15 Jun 2025 · edited#108

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

0 likes 13mo
CR
c.ramosTL2 Moderator16 Jun 2025#109

Sample matrix effects: if a sample is dissolved in a complex matrix, other compounds in the matrix can suppress the peptide signal. Clean samples give higher sensitivity than dirty samples.

2 likes 13mo
CB
c.bakkerTL2 Moderator17 Jun 2025#110

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

0 likes 13mo
D
DKwiatkowskiTL3Regular18 Jun 2025#111

post #110 is right about the mechanism and I think understates the practical bit.

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

31 likes 13mo
JL
j.lokkenTL2 Moderator19 Jun 2025#112

Worth separating two things that post #108 runs together.

Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity.

0 likes 13mo
SS
s.stavrianosTL2Member20 Jun 2025#113
Okafor, post #70: On post #66 — agreed on the reasoning, with one qualification. Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. Go to post

Thank you for the correction. I have edited my earlier post with a note rather than silently, so the thread still makes sense to read. The error was mine and it was the kind that comes from remembering a figure instead of looking it up.

6 likes in reply to #70 13mo
AV
a.villalobosTL2 Moderator21 Jun 2025 · edited#114

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

15 likes 13mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Coming back to: Oxidation and deamidation mass shifts, tabulated
Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable. Working through the identity arithmetic and I would like it checked. retatrutide has a…
HFSLADAKTI+34 38 4.8k 10mo
Coming back to: Deconvolution of a multiply charged envelope
On the subject in the title: Deconvolution of a multiply charged envelope Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
VRCVEL 2 12k 22mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
RMLSMHMINM+121 129 25k 17mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo
Distinguishing a deletion sequence by mass alone — the long version
On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion. I would like to understand what this number means before I repeat it…
TFVKPFIDLF+38 44 60k 11mo

Related topics — sharing the tags counter-ion content, mass error (ppm), charge states

TopicParticipantsRepliesViewsActivity
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h
Residual solvents and scavengers: what is looked for
On the subject in the title: Residual solvents and scavengers: what is looked for Working notes rather than a conclusion. Posting the method first, because I know what the first three replies will otherwise…
DOEVABKPLG+19 23 39k 14mo
Aggregates and why a purity assay may miss them entirely
Aggregates and why a purity assay may miss them entirely Writing it up because I had to work it out twice and would rather nobody else did. A documentation question rather than an analytical one. I have a…
RVIDSSVKFP+61 69 2.9k 6mo
Follow-up: Trifluoroacetate content and its consequences
On the subject in the title: Trifluoroacetate content and its consequences Working notes rather than a conclusion. Posting the method first, because I know what the first three replies will otherwise be.…
AVABAZDTRG+12 16 5.7k 21mo
Attributing a new peak: degradation product or excipient?
Asking directly, because I could not find a straight answer: Attributing a new peak: degradation product or excipient? Posting the method first, because I know what the first three replies will otherwise be.…
SHOWM 2 15k 18mo