Analytics · Method validation · continued

Repeatability versus intermediate precision posts 121–135

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

CB
c.bakkerTL2 Moderator2 Jul 2025#121

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

0 likes 13mo
MB
m.brobergTL2 Moderator2 Jul 2025#122

Worth separating two things that post #118 runs together.

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

3 likes 13mo
SL
s.leclercTL4 Moderator2 Jul 2025#123
Staff post. Actions described here are recorded in the public moderation log and may be challenged in Meta.

This follows post #120 rather than contradicting it.

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

11 likes 13mo
MI
m.ibarraTL2 Moderator2 Jul 2025#124
KTurkington, post #33: For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use. Go to post

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

24 likes in reply to #33 13mo
AF
a.friskTL2 Moderator2 Jul 2025 · edited#125

post #124 answers the question as asked. The question underneath it is different.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

0 likes 13mo
PM
p.mbekiTL22 Jul 2025#126
NM
n.moreauTL2 Moderator3 Jul 2025#127

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

7 likes 13mo
JN
j.nascimentoTL2 Moderator3 Jul 2025#128
unit_conversion, post #47: Coming back to post #45, because the follow-up matters more than the original answer. Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal… Go to post

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

17 likes in reply to #47 13mo
TH
TL4_HalvorsenTL4Leader · Journal club3 Jul 2025#129
a.frisk, post #36: On post #32 — agreed on the reasoning, with one qualification. Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically. Go to post

For anyone arriving from a search: the marked solution above is the direct answer, and the replies underneath it add the caveats that make it safe to use.

3 likes in reply to #36 13mo
CA
c.amankwahTL2 Moderator3 Jul 2025#130

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

11 likes 13mo
RV
r.venkatesanTL3Wiki editor3 Jul 2025#131
unit_conversion, post #47: Coming back to post #45, because the follow-up matters more than the original answer. Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal… Go to post

Coming back to post #129, because the follow-up matters more than the original answer.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

0 likes in reply to #47 13mo
KP
k.pereiraTL2 Moderator3 Jul 2025#132
b.oyler, post #18: Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples. Go to post

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

23 likes in reply to #18 13mo
MM
maintenance_modeTL3Regular3 Jul 2025 · edited#133

Two things before anyone answers the substance.

First, the context in the first post is clear and specific. Second, the question is framed so that an answer can actually address it. Both are the norm here and both matter more than they sound.

10 likes 13mo
AP
au.pereiraTL2 Moderator3 Jul 2025#134

System suitability: injections run at the start of a batch to establish that the instrument and column are performing. Acceptance criteria typically include replicate precision (RSD ≤2%), peak tailing (0.8–1.5), theoretical plates (>2000), and resolution (>1.5).

3 likes 13mo
TY
two_year_lineTL3Regular3 Jul 2025#135
s.bergstrom, post #34: Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

I read post #133 twice before replying, because I had assumed the opposite.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

32 likes in reply to #34 13mo

Suggested topics

TopicParticipantsRepliesViewsActivity
About the Method validation category
Specificity, linearity, precision, accuracy, robustness, system suitability, and why two labs disagree. This post is a community wiki: any member at trust level 3 or above can edit it, and every edit is…
KTVPAKAJV+1 5 468 5mo
When to suspect the method rather than the sample
When to suspect the method rather than the sample — that is the question, and I have not found it answered plainly anywhere I have looked. Posting the method first, because I know what the first three replies…
MMCCYMRMBJ+109 115 63k 15mo
Second pass at: Robustness: the parameters worth deliberately varying
On the subject in the title: Second pass at: Robustness: the parameters worth deliberately varying Working notes rather than a conclusion. I would like to understand what this number means before I repeat it…
BSFWDSCVOO+61 66 9.5k 3mo
Revisiting: Spike-and-recover accuracy for a peptide assay
Revisiting: Spike-and-recover accuracy for a peptide assay — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it anywhere.…
SCVKEC 2 6.8k 21h
What a stability-indicating method is and how you demonstrate one
Asking directly, because I could not find a straight answer: What a stability-indicating method is and how you demonstrate one Working through the identity arithmetic and I would like it checked. retatrutide…
EFRJPKISFD 4 57k 6mo

Related topics — sharing the tags precision & repeatability, recovery, LOD & LOQ

TopicParticipantsRepliesViewsActivity
Repeatability versus intermediate precision — what changed since
Posting this under the heading it deserves: Repeatability versus intermediate precision — what changed since Everything below is what sits behind that. Working through the identity arithmetic and I would like…
DOBRANTIFN 4 21k 5h
Coming back to: Where the integrator drew the baseline, and how much it moved the number
On the subject in the title: Where the integrator drew the baseline, and how much it moved the number Working notes rather than a conclusion. Posting the method first, because I know what the first three…
AHNMTNAFMB+20 24 2.7k 10mo
Why quantitation is the hard part, not detection
Why quantitation is the hard part, not detection I have a specific reason for asking rather than idle curiosity, and the context is below. Posting the method first, because I know what the first three replies…
APWNHESTJF+11 15 27k 12mo
Relative retention time as a better inter-laboratory comparator — does this still hold?
Relative retention time as a better inter-laboratory comparator — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below. A documentation question…
EARFWNFWOO+11 15 21k 21mo
Coming back to: What consumer-accessible tests can and cannot determine
Asking directly, because I could not find a straight answer: What consumer-accessible tests can and cannot determine Posting the method first, because I know what the first three replies will otherwise be.…
PKNWVNHK+4 8 25k 22mo